Jump to content

Trouble with a couple of CompaqDaq Modules


UnderDog

Recommended Posts

Hi,

This is the error from Signal Express:

Error -200077 occurred at DAQ Assistant

Possible Reason(s):

Requested value is not a supported value for this property.

Property: SampTimingType

You Have Requested: On Demand

You Can Select: Sample Clock

Device: cDAQ12

You cannot use this device in this configuration. In LabVIEW you can use the "Set Clock" vi to use something like the "OnboardClock." Although I guess this is not true "Sample on demand", however, the "Get 1 Sample" function gets only the last value. (So it looks like on demand sampling)

Link to comment

Join the conversation

You can post now and register later. If you have an account, sign in now to post with your account.

Guest
Unfortunately, your content contains terms that we do not allow. Please edit your content to remove the highlighted words below.
Reply to this topic...

×   Pasted as rich text.   Paste as plain text instead

  Only 75 emoji are allowed.

×   Your link has been automatically embedded.   Display as a link instead

×   Your previous content has been restored.   Clear editor

×   You cannot paste images directly. Upload or insert images from URL.

×
×
  • Create New...

Important Information

By using this site, you agree to our Terms of Use.