BIB Posted April 28, 2016 Report Share Posted April 28, 2016 Hello , I am working with a labview program which hardware is composed from a Pickering card 40-535-021 and a keithley 2400 . i am using visa for communication The application used to work on the older chassis (NIPXI 1042Q )but when we took a newer one(NI PXI 1031) this problem happened. When the Keithley is on the program can't find the pickering card but when the keithley is off, it can find it . there is a conflict between the two . it like the program does not recognize the Pickering card properly. does any body have an idea ? Any help is welcome ! regards BIB Quote Link to comment
Bryan Posted April 28, 2016 Report Share Posted April 28, 2016 Look to see if the Pickering card is being detected at a different address/location by the processor when the Keithley is ON vs when it's OFF. You should be able to do it either via NI MAX or a soft front panel of some kind. Quote Link to comment