QUOTE (ASTDan @ May 7 2008, 06:50 AM)

I can't address your question specifically because I have not dealt with this hardware, but I can speak in general terms...

When dealing with any temperature measurement device you have to worry about accuracy and repeatability. A published spec only says what the unit is required to do, but most of the time the actual performance is better than the spec. You can calibrate out errors in accuracy by measuring a series of known source temperatures (or by comparing measurements of multiple sources with measurements from a high accuracy device) and generating a lookup table that shows the offset from the actual temperature to the measured temperature.

We generally do this by taking 5-10 measurements about 10 seconds apart and averaging them. If we are measuring a 200°C source and our average measurement is 200.7°C then at 200°C we have an offset of -0.7°C. Repeat this process different temperatures until you can generate a relatively good offset curve. Then any time you get a measurement from your device, add the offset at that temperature to give you a more accurate result.

Repeatability is a bit trickier. Really all you can do is characterize your device and hope that it is better than the published spec. We generally find that on high quality hardware the repeatability is much better than the published spec. Taking an average of measurements over time can also help to minimize the damage done by repeatability error.

Occasionally you should recheck your device against a known source to ensure your offset table is still accurate. Calibrations can drift over time. There are also labs that will do the characterization for you and generate the offset table (we send our temperature standards to NIST for calibration and then use those to calibrate our units that don't require such high accuracy) but it can be expensive.

-Toby

Link to comment

QUOTE (ASTDan @ May 7 2008, 09:50 AM)

The accuracy of the measuring device can only be as accurate as the transducer. Here's a table of TC accuracies I found: http://www.microlink.co.uk/tctable.html. Omega has simlar charts based on the number of coefficients you use to calculate the polynomial.

In our specs, we usually use an accuracy of +- 3 deg C for TCs in general regardless of how we measure them. Also, TCs can drift over time depending upon what the operating temp is and oxidation levels. We've seen a 5 deg C drift for type K thermocouples used in high temps over the course 2-3 years.

For situtations where we need better accuracy, we use precision RTDs.

Link to comment

Join the conversation

You can post now and register later. If you have an account, sign in now to post with your account.

Guest
Reply to this topic...

×   Pasted as rich text.   Paste as plain text instead

  Only 75 emoji are allowed.

×   Your link has been automatically embedded.   Display as a link instead

×   Your previous content has been restored.   Clear editor

×   You cannot paste images directly. Upload or insert images from URL.

×
×
  • Create New...

Important Information

By using this site, you agree to our Terms of Use.