Hi,

I would like to ask if any have experiment with Omega load cell LCMFL series? By some reasons, I cannot get any response from the load cell. So I need help.

In my project I have a load cell LCMFL 20N connect with NI 9237 card. I set it up with full bridge (verify with customer service), internal, 5 V excitation. The scale is 10N 1mV/V. I cannot get any response from it.

Then I tried to connect it with a 5VDC power supply and use multimeter to measure but the response signal is not changed either. So I wonder if I do anything wrong with the wiring or not.

 

Regards,

Thang Nguyen

Edited by Thang Nguyen
Link to comment

Hello Thang,

 

The documentation found here: http://www.omega.com/Pressure/pdf/LCMFL.pdf seems to indicate that these load cells require exitation voltage on the Red and Black wires.  This is typical of most load cells.

 

I can't say for sure why your app is not working.  Have a look at the following documentation and check your wiring.

 

http://digital.ni.com/public.nsf/allkb/D02125BF0BFC6130862576650001F14C

http://sine.ni.com/psp/app/doc/p/id/psp-168/lang/en

Link to comment
  • 2 weeks later...

Join the conversation

You can post now and register later. If you have an account, sign in now to post with your account.

Guest
Unfortunately, your content contains terms that we do not allow. Please edit your content to remove the highlighted words below.
Reply to this topic...

×   Pasted as rich text.   Paste as plain text instead

  Only 75 emoji are allowed.

×   Your link has been automatically embedded.   Display as a link instead

×   Your previous content has been restored.   Clear editor

×   You cannot paste images directly. Upload or insert images from URL.

×
×
  • Create New...

Important Information

By using this site, you agree to our Terms of Use.