Software architecture for modular measurement system in a battery and electronics testing laboratory
Greetings,
To set the scene, our University wants to build a laboratory where we can test batteries and electronic devices with our equipment. Just to list a few of them we have:
- Climate chamber
- cRIO-9039 for data acquisition and safety control
- Electronic load
- Multiple power supplies
- Multimeters
- Oscilloscope
- LCR meter
- EIS measurement instrument
and so on....
All of these devices can be interfaced through LabVIEW, and I am certain that this list will increase in the future.
The main concept:
- We have a sotfware, where we can create the test sequence, step by step (i.e.: Configure powersupply, delay(ms), enable ps. output, delay(ms), disable output ). The output would be a recipe file, which include each step, and its parameters.
- The main software, where you can import the test sequence, and execute it. The sequences can include multiple measurement instruments. The data logging is continuous.  If there is an emergency, the measurement is aborted.
   The imported sequence file would be executed step by step in a case structure (or if there is any other possibility).
My question is: What kinf of architecture would be suitable for the main software? I have already developed larger applications using QMH, but I have experienced its limitations. Im also looking at OOP concpets, but I have no experience with i so far.
I know that TestStand is similar to our needs, but currently we dont have a budget to purchase it.
Thank you for your help
First, I'm in the same boat as you are, and here are few of my insights:
You are trying to create a test sequencer - There are off the shelf products that do that (Teststand is NI's solution - $$$). For architecture, anything from the basic QMH to a full blown actor framework with asynchronous actors can be used - in my case, I have used QMH with few extra threads for safety, logging and report generation.
OOP is a way for the designers to think about software modules and abstraction. Do you need such things? Are you planning on supporting multiple models of PSU, loads and other test instruments? If so, OOP provides a convenient way to create those abstraction layers. If you decide on using classes, you should avoid using PPL, they break each other in weird and wonderful ways.
For data logging, I suggest both a central parametric database and local Excel or Calc reports.
Test templates and deployment: You should consider creating a standardized test template project, along with build specifications, so your operators will use an executable and not the full LV IDE.
  • 1 month later...
I'm prone to use JKI state machines for this sort of simple test sequencing. It is easy to learn, and surprisingly powerful once you become proficient with it. You can also use JKI State Machine Objects (traditional JKI SM wrapped in a class) which makes it simple to support multiple parallel state machines, and which also supports using events.
Regards,
Rick
OK. This is how I design systems like this.
1. TCP. Each subsystem has a TCP interface. This allows spinning up "instances" then connecting to them even across networks. You can rationalise the TCP API and I usually use a thin LabVIEW wrapper around SCPI (most of your devices will support SCPI). You can also use it to make non SCPI compliant devices into SCPI ones (your Environmental Chamber - EC - is probably one that doesn't support SCPI). If you do this right, you can even script entire tests from a text file just using SCPI commands.
2. Services. Each subsystem offers services. These are for when #1 isn't enough and we need state. A good example of this is your Environmental Chamber. It is likely you will have temperature profiles that  control signals and measurements need to be synchronized with. While services may be devices (a DVM, for example), services can also be synchronization logic that sequences multiple devices. If you put that logic in your EC code, it will fix that code for that specific sequence so don't do that. Instead  use services to glue other devices (like the DVM and EC) into synchronization processes. Along with #1, this will form the basis of recipe's that can incorporate complex state and sequencing. In this way you will compartmentalize your system into reusable modules. First thing you should do is make a "Logging" service. Then when your devices error they can report errors for your diagnostics. The second thing should be a service that "views" the log in real-time so you can see what's going on, when it's going on. (This is why we have logging levels).
3. Global State. If you have 1 & 2 this can be anything. It can be a text file with a list of SCPI commands (#1). It can be a service you wrote in #2, Test Stand, web page or a bash/batch script. This is where you use your recipe's to fulfill a test requirement.
4. You will need to think carefully about how the subsystems talk to each other. For example. Using SCPI a MEAS :VOLT:DC? command returns almost instantly (command-response pattern). However for the EC you may want to wait until  a particular temperature has been reached before issuing MEAS :VOLT:DC?. The problem here is that SCPI is command-response but the behavior required is event driven. One could make the the TCP interface (#1) of the EC accept MEAS :TEMP? where the command doesn't return unless the target setpoint has been reached. However, this won't work reliably and requires internal state and checks for the edge cases though. So it may actually be better in #2. There are a number of ways to address these things using #1, #2 or #3 and that is why you are getting the big bucks.
You will notice I haven't mentioned specific technologies here (apart for TCP). For #1 you shouldn't need anything other than reentrant VI's and VI Server. For #2 you can use your favourite foot-shooting method but notice that you are not limited to one type and can choose an architecture for the specific task (they don't all have to be QMH, for example). For #3 you don't even have to use LabVIEW.
Edited by ShaunR
Nicely described approach Shaun.  I am doing something similar with a lab automation project that involves vacuum chamber and multi-zone temperature control.  Elected to use the messenger framework, which supports 'spinning up' instances and TCP capability for remote devices like cRIO.  The messenger architecture also supports many types of asynchronous messaging beyond simple synchronous command/response.  Your idea of decoupling EC sequencing logic is good, moving it out of the EC subsystem to allow synchronization with other subsystems during ramp/soak profiles.   Hey remember Test Stand Lite?  This is where an such a scripting component would really shine and be a great benefit to the community.
Edited by viSci
Expand
You can script with text files with #1. Just adding a feature to delay N ms gets you most of the way to a full scripting language (conditionals and for-loops are what's left but a harder proposition). However, for a more general solution I use services in #2  with queue's for inputs and events for outputs. There was a discussion ages ago at whether queues were needed at all since LabVIEW events have a hidden queue but you can't push to the front of them (STOP message ;)) and, at the time, you couldn't clear them so I opted for proper queues. So, architecturally, I use "many to one" for inputs and "one to many" for outputs.
I work in a battery test lab and was the architect for the testing platform we use.  Our main hardware is an embedded cDAQ, or cRIO running Linux RT.  This is the hardware that runs the actual sequence, and talks to the various hardware.  We use the RT platform more for reliability, and less for its determinism.  Timing is of course important, but we don't have any timed loops. mS precision is really all we care about. The architecture is built around User Events with asynchronous loops dedicated to specific tasks.  These events can be triggered from the RT application, or from a device on the network using Network Streams. A low level TCP would probably be better, but the main sequence itself isn't sent, step by step from Windows. Instead the sequence file is downloaded via WebDAV, then told to run it. The RT then reads the file and performs the step one at a time. Windows from this point is just for monitoring status, and processing logs. This is important to us since Windows PCs might restart, or update, or blue screen, or have any other number of weird situations, and we wanted our test to just keep going along.  This design does mean that any external devices that use Windows DLLs for the communication can't be used. But network controlled power supplies, loads, cyclers, chambers, and chillers all are fine. Same with RS-232/485.  Linux RT supports VISA, and we plugged in a single USB cable to a device that gives us 8 serial ports.  If money were no object I suppose we would have gone with PXI but it is pretty over kill for us.
On the Windows side we do also have a sequence editor.  The main reason we didn't go with TestStand is because we want that sequence to run entirely on Linux RT.  Because the software communicates over User Events, and Network Streams sending User Events, we can configure the software to run entirely on Windows if we want.  This of course takes out many of the safety things I talked about, but to run a laptop to gather some data, control a chamber, battery, cycler, or chiller for a short test is very valuable for us.
As for the design, we don't use classes everywhere. Mostly just for hardware abstraction.  The main application is broken up into Libraries, but not Classes.  If I were to start over maybe I would use classes just to have that private data for each parallel loop, but honestly is isn't important and a type def'd cluster works fine for the situation we use it in.
  On 5/6/2025 at 1:25 PM, ShaunR said:
How does this work?
Expand
I have a network stream actor (not NI's actor but whatever) that sits and handles the back and forth. When you want work to happen like "Set PSU Output" you can state which instance you are asking it to (because the actor is reentrant), and which network location you want. The same VI is called, and can send the user event to the local instance, or will send a user event to the Network Stream loop, which will send the request for a user event to be ran on the remote system, and then reverse it to send the reply back if there is one.  I like the flexibility of having the "Set PSU Output" being the same VI I call if I am running locally, or sending the request to be done remotely.  So when I talked about running a sequence, it is the same VIs called, just having its destination settings set appropriately.
Expand
Yes. It is the "send user event" that I'm having difficulty with. User events are always local and require a prototype so how do you serialise a user event to send it over a stream?
Expand
Variants and type defs.  There is a type def for the request, and a type def for the reply, along with the 3 VIs for performing the request, converting the request, and sending the response.  All generated with scripting along with the case to handle it.  Because all User Events are the same data type, they can be registered in an array at once, like a publisher/subscriber model.  Very useful for debugging since a single location can register for all events and you can see what the traffic is.  There is a state in a state machine for receiving each request for work to be done, and in there is the scripted VI for handling the conversion from variant back to the type def, and then type def back to variant for the reply.
When you perform a remote request, instead of sending the user event to the Power Supply actor, it gets sent to the Network Streams actor.  This will get the User Event, then send the data as a network stream, along with some other house keeping data to the remote system.  The remote system has its Network Stream actor running and will get it, then it will pull out the data, and send the User Event, to its own Power Supply actor.  That actor will do work, then send a user event back as the reply.  The remote Network Stream actor gets this, then sends the data back to the host using a Network Stream. Now my local Network Stream actor gets it, and generates the user event as the reply.  The reason for the complicated nature, is it makes using it very simple.
Expand
IC. so you have created a cloning mechanism of User Events - reconstruct pre-defined User Event primitives locally with the data sent over the stream?
Expand
Yes. The transport mechanism could have been anything, and as I mentioned I probably should have gone with pure TCP but it was the quickest way to get it working.
Expand
LabVIEW network streams are fine. I wouldn't worry about the transport too much. Network Streams have a nice way of integrating into API's.
What I was looking at were network events where you don't have to synchronise prototypes throughout a system on different machines (changing an event prototype usually breaks an Event  Structure). Everything for my events is a string so this is not a problem but it makes parsing tricky. This was one of the reasons I wanted "Named Events" (events can be named like queues can) but they botched the downstream polymorphism. I was wondering whether you had found an elegant way of serialising events (a bit like protocol buffers but without the compiler).
Edited by ShaunR
Yeah I tried making it as elegant as possible, but as you said there are limitations, especially with data type propagation.  I was hoping to use XNodes, or VIMs to help with this, but in practice it just made things overly difficult.  I do occasionally get variant to data conversion issues, if say the prototype of a request changes, but it didn't get updated on the remote system. But since I only work in LabVIEW, and since I control the release of all the builds, it is fairly manageable.  Sometimes to avoid this I will make a new event entirely, to not break backwards compatibility with older systems, or I may write version mutation code, but this has performance hits that I'd rather not have. Like you said, not always elegant.
Expand
Have you played with scripting event prototypes and handlers from JSON strings?
Expand
No but that is a great suggestion to think about for future improvements. At the moment I could do the reverse though. Given the Request/Reply type defs, generate the JSON strings describing the prototypes.  Then replacing the Network Streams with HTTP, or TCP could mean other applications could more easily control these remote systems.
Expand
That's the easy bit. It's much easier getting stuff into other forms than it is reconstituting them because we can't create objects and primitives. This is why JSON libraries have very straightforward encoders that can take any type but you have all sorts of awkward VI's for getting them out into LabVIEW again.
If you are going to use JSON strings you might as well not use LabVIEW types at all ;). Add the Network Stream endpoint to it and you're good to go. Getting it back out again is where you will find the problems unless, of course, the device uses strings too (SCPI)
The popular serializer/deserializer problem. The serializer is never really the hard part (it can be laborious if you have to handle many data types but it's doable) but the deserializer gets almost always tricky. Every serious programmer gets into this problem at some point, and many spend countless hours to write the perfect serializer/deserializer library, only to abandon their own creation after a few attempts to shoehorn it into other applications. 🙂
Edited by Rolf Kalbermatter
Expand
We are on a hiding to nothing as we can't create objects. I abandoned those thoughts over a decade ago (maybe even 2 decades ago :ph34r:). It is feasible with scripting but so slow and won't work in built applications. The half-way house is to use scripting to make the labview prototypes for typdefs and handlers then load dynamically as plugins but it's a lot of infrastructure just to propagate what are hopefully rare changes. I did play with that (hence my question to hooovahh) but in the end went for a string based solution to avoid it altogether.
What happened to AQ's behemoth of a serializer? Did he ever get that working?
Edited by ShaunR
Expand
I believe to have read somewhere that he eventually conceded that it was basically an unsolvable problem for a generic, easy to use solution.
But that may be my mental interpretation of something he said in the past and he may not agree with that.